04-01-08, 03:05 AM
Hi,
I just purchased an eye-one display 2 and encountered a strange problem with the software out of the box. Everything seemed alright until i copied the eyeone.dll (2.0.0.4) to the hcfr directory and started doing grayscale measurements with hcfr 2.0.1.
Up to a certain luminance level everything looked alright then the readings went haywire for an IRE Step or two and then everything went back to normal for higher IREs. It claimed there was no red (0% !)at all between 15<Y<25 which was visibly wrong as such a strong cyan cast cannot be missed.
This was totally independent of testpatterns source, direct/indirect measurements or position of the probe. If i put the projector in high lamp mode i could shift the problem to lower IREs which read fine on the lower lamp mode before.
At first it seemed to me the probe was simply faulty and had a problem at a certain luminance level. I was about to send the probe back when i tried one more thing - updating the software from the site to colormatch 3.6.2. and using the new eyeone.dll (3.4.0.131).
With the new eyeone.dll all problems were gone and it read the grayscale about 3 times faster then before. I did about 10 readings from different positions just to make sure (the readings outside the problem luminance were about the same as before). I hope this wasn't just luck and if its a real solution, maybe this will save others some time.
I am still curious if this is a known issue and if somebody else had this problem before?
regards,
jojodyne
I just purchased an eye-one display 2 and encountered a strange problem with the software out of the box. Everything seemed alright until i copied the eyeone.dll (2.0.0.4) to the hcfr directory and started doing grayscale measurements with hcfr 2.0.1.
Up to a certain luminance level everything looked alright then the readings went haywire for an IRE Step or two and then everything went back to normal for higher IREs. It claimed there was no red (0% !)at all between 15<Y<25 which was visibly wrong as such a strong cyan cast cannot be missed.
This was totally independent of testpatterns source, direct/indirect measurements or position of the probe. If i put the projector in high lamp mode i could shift the problem to lower IREs which read fine on the lower lamp mode before.
At first it seemed to me the probe was simply faulty and had a problem at a certain luminance level. I was about to send the probe back when i tried one more thing - updating the software from the site to colormatch 3.6.2. and using the new eyeone.dll (3.4.0.131).
With the new eyeone.dll all problems were gone and it read the grayscale about 3 times faster then before. I did about 10 readings from different positions just to make sure (the readings outside the problem luminance were about the same as before). I hope this wasn't just luck and if its a real solution, maybe this will save others some time.
I am still curious if this is a known issue and if somebody else had this problem before?
regards,
jojodyne