post #1 of 82 Old 10-24-2013, 07:06 PM - Thread Starter
Member
 
jrodefeld's Avatar
 
Join Date: Sep 2007
Posts: 79
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 4 Post(s)
Liked: 11
Hello,

I bought a used Eye One Pro rev D on Ebay six months ago. I think it would be wise to get it recalibrate as I don't think it has been certified in probably four or more years. And I have no idea how it was stored or treated by its previous owner.

I am not doing calibrations for anyone but myself and family. I just need to know that the readings are accurate. But Xrite says they will charge me like $250 for recertification and if it needs any adjustments it will cost over $400 total.

That is just too much money at this time. I want to know if there are any low cost alternatives? Couldn't another amateur with a recently certified reference meter be able to compare the two and tell me how accurate my meter is?

And if is is found to be inaccurate even if they could not actually fix the mechanical parts of the meter couldn't they just give me the values to plug in as offsets so I would get accurate readings?

If anyone can do anything like this I would pay you for the effort. Please let me know what my options are.

Thanks.
jrodefeld is offline  
Sponsored Links
Advertisement
 
post #2 of 82 Old 10-24-2013, 08:15 PM
Advanced Member
 
CalWldLif's Avatar
 
Join Date: Feb 2011
Location: South Bay'Los Angeles California.
Posts: 518
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 70 Post(s)
Liked: 29
I had/have the same desire to "prove" my i1pro rev D.

My phone call to x-rite was quoted as 400 to start,

Loving D65
CalWldLif is offline  
post #3 of 82 Old 10-24-2013, 08:25 PM
AVS Special Member
 
turbe's Avatar
 
Join Date: Feb 2001
Location: Lake Tahoe, NV
Posts: 4,437
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 7 Post(s)
Liked: 90
It's a waste of money for enthusiasts.. I have never seen a report of an undamaged i1Pro Rev D failing re-certification and I have been tracking this for years now.. this meter is still used by Professionals globally, they do re-certify, as all Professionals should do.

Alternatives to recertification
If you want it checked, I suggest you hire a Professional with a JETI, Photo Research PR-655 (with 5nm option), Photo PR-670 or equivalent reference Spectro (or perhaps even one with a recently certified i1Pro / i1Pro2) to come to your house for the verification and possibly a full calibration.. it can be a great educational experience if you get a Calibrator that also takes the time to explain the process while the calibration is happening on your actual display.

.

Need to find a Professional Calibrator? Click Here to PM me with your Display & City

Calibrator List - Pioneer ISFccc Interface

Calibration Reports - Pioneer

 

ControlCAL™
Designed by Calibrators for Calibrators™

No need to fumble through the Display's Menu with its Remote Control™

turbe is offline  
post #4 of 82 Old 10-24-2013, 08:44 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
One thing you could do is purchase some light bulbs/LED bulbs that have very well specified spectra, and measure them with the i1 pro. You can also do the same in reflectance mode by measuring standard color patches (munsell I think is a good one).
spacediver is offline  
post #5 of 82 Old 10-24-2013, 08:55 PM
AVS Special Member
 
turbe's Avatar
 
Join Date: Feb 2001
Location: Lake Tahoe, NV
Posts: 4,437
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 7 Post(s)
Liked: 90
the bulb method would not work (even those bulbs, at least available known options to enthusiasts/enthusiasts budget, are off more than the meter's spec'd tolerance) and the munsell patches wouldn't be acceptable in this case IMO since the meter's lamp and the software (for our purpose) comes into play.

.

Need to find a Professional Calibrator? Click Here to PM me with your Display & City

Calibrator List - Pioneer ISFccc Interface

Calibration Reports - Pioneer

 

ControlCAL™
Designed by Calibrators for Calibrators™

No need to fumble through the Display's Menu with its Remote Control™

turbe is offline  
post #6 of 82 Old 10-24-2013, 09:20 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
I'm guessing it's because reflectance mode would eliminate a lot of the optics pipeline, right? (including issues with stray light). My understanding is that reflectance mode has a higher tolerance for bandpass errors than emmissive mode, so errors that wouldn't show up in reflectance mode would show up when measuring light sources.


What about sunlight - are there any conditions under which the spectra of sunlight is reliably constant (say during a summer solstice in a particular geographic location on a cloudless day at noon?) Would it even be possible to measure solar radiance with an i1 pro? Even if it were possible, the sun's spectral distribution is very smooth, and would not challenge a spectroradiometer in the same way that a more spiky source would. Is it really that hard to find affordable LEDs with tolerances tighter than the i1 pros accepted tolerances?
spacediver is offline  
post #7 of 82 Old 10-24-2013, 09:48 PM
AVS Special Member
 
turbe's Avatar
 
Join Date: Feb 2001
Location: Lake Tahoe, NV
Posts: 4,437
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 7 Post(s)
Liked: 90
not to mention how would we align an i1Pro(1/2) and an reference meter with that in an controlled environment, let alone an enthusiast's..

Need to find a Professional Calibrator? Click Here to PM me with your Display & City

Calibrator List - Pioneer ISFccc Interface

Calibration Reports - Pioneer

 

ControlCAL™
Designed by Calibrators for Calibrators™

No need to fumble through the Display's Menu with its Remote Control™

turbe is offline  
post #8 of 82 Old 10-24-2013, 10:12 PM
AVS Special Member
 
Chad B's Avatar
 
Join Date: Oct 2002
Location: Piqua, OH
Posts: 2,158
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 127 Post(s)
Liked: 539
Quote:
Originally Posted by jrodefeld View Post

Couldn't another amateur with a recently certified reference meter be able to compare the two and tell me how accurate my meter is?

And if is is found to be inaccurate even if they could not actually fix the mechanical parts of the meter couldn't they just give me the values to plug in as offsets so I would get accurate readings?

If anyone can do anything like this I would pay you for the effort. Please let me know what my options are.

Thanks.
Where are you located?

ISF/THX calibrator with Jeti 1211 reference spectro
Latest reviews:Samsung UN65HU8550UN65HU9000, UN85HU8550, LG 55EC9300, LG 55EA8800 (including 3D)

Why I don't publish settings, LG OLED calibration tips, Review index and rankings, Florida tour Jan '15, Texas tour March '15
Chad B is online now  
post #9 of 82 Old 10-24-2013, 10:29 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Quote:
Originally Posted by turbe View Post

not to mention how would we align an i1Pro(1/2) and an reference meter with that in an controlled environment, let alone an enthusiast's..


well you wouldn't need a reference meter, right? Assuming you have confidence in the SPD /chromaticity of your light source, that IS your "reference source".
spacediver is offline  
post #10 of 82 Old 10-24-2013, 10:30 PM
AVS Special Member
 
turbe's Avatar
 
Join Date: Feb 2001
Location: Lake Tahoe, NV
Posts: 4,437
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 7 Post(s)
Liked: 90
I'm glad Chad posted, he has a JETI 1211 and it's most likely he also has the software (for our purpose) you use.

Need to find a Professional Calibrator? Click Here to PM me with your Display & City

Calibrator List - Pioneer ISFccc Interface

Calibration Reports - Pioneer

 

ControlCAL™
Designed by Calibrators for Calibrators™

No need to fumble through the Display's Menu with its Remote Control™

turbe is offline  
post #11 of 82 Old 10-24-2013, 10:43 PM
AVS Special Member
 
turbe's Avatar
 
Join Date: Feb 2001
Location: Lake Tahoe, NV
Posts: 4,437
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 7 Post(s)
Liked: 90
Quote:
Originally Posted by spacediver View Post

well you wouldn't need a reference meter, right? Assuming you have confidence in the SPD /chromaticity of your light source, that IS your "reference source".

which light source and in regards to our purpose of display calibration?

'confidence' is the key, isn't it?

Need to find a Professional Calibrator? Click Here to PM me with your Display & City

Calibrator List - Pioneer ISFccc Interface

Calibration Reports - Pioneer

 

ControlCAL™
Designed by Calibrators for Calibrators™

No need to fumble through the Display's Menu with its Remote Control™

turbe is offline  
post #12 of 82 Old 10-24-2013, 10:58 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
well that's why I was asking about the sun, or some LED's with tight tolerances.
spacediver is offline  
post #13 of 82 Old 10-24-2013, 11:07 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
This one for example. Might be interesting to request a quote just to get an idea.
spacediver is offline  
post #14 of 82 Old 10-25-2013, 03:55 AM - Thread Starter
Member
 
jrodefeld's Avatar
 
Join Date: Sep 2007
Posts: 79
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 4 Post(s)
Liked: 11
Quote:
Originally Posted by Chad B View Post

Where are you located?

I'm in Southern California close to Santa Barbara.
jrodefeld is offline  
post #15 of 82 Old 10-25-2013, 07:06 AM
AVS Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 2,096
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 45 Post(s)
Liked: 72
Quote:
Originally Posted by spacediver View Post

This one for example. Might be interesting to request a quote just to get an idea.

No colorimetry specs. Worthless in that case. In any event, spec'ed or not, you'd still have to measure it with a reference instrument to know what your unit under test should be reporting. Otherwise, you're guessing.

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931
Rolls-Royce is online now  
post #16 of 82 Old 10-25-2013, 09:44 AM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Quote:
Originally Posted by Rolls-Royce View Post

No colorimetry specs. Worthless in that case. In any event, spec'ed or not, you'd still have to measure it with a reference instrument to know what your unit under test should be reporting. Otherwise, you're guessing.

If you had colorimetry specs, and you trusted the source, why would you need a reference instrument?
spacediver is offline  
post #17 of 82 Old 10-25-2013, 11:53 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,518
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 169 Post(s)
Liked: 562
Find a DIYer than lives near to your place with a i1PRO/i1PRO2 that it's under a it's valid certificated period and compare the spectral response of the 2 meters.

There is no other way to find out, unless you send it to germany, to X-Rite's repair central to re-certificate your meter by tweaking it internally, only the factory can access it internally.

But the cost is very high for that service....

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is online now  
post #18 of 82 Old 10-25-2013, 12:26 PM
Advanced Member
 
ElectronicTonic's Avatar
 
Join Date: Feb 2008
Location: Ventura, CA
Posts: 981
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 21 Post(s)
Liked: 68
Quote:
Originally Posted by jrodefeld View Post

I'm in Southern California close to Santa Barbara.
I sent you a PM.
ElectronicTonic is offline  
post #19 of 82 Old 10-25-2013, 01:50 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Quote:
Originally Posted by Rolls-Royce View Post

No colorimetry specs. Worthless in that case. In any event, spec'ed or not, you'd still have to measure it with a reference instrument to know what your unit under test should be reporting. Otherwise, you're guessing.

I've just sent a request for further information, including pricing. Will report back if and when I hear from them.

If it turns out that the colorimetry specs are included, and are stable and accurate, then I don't understand why you couldn't use a set of these to test your instruments. Why on earth would you need a reference instrument, when you have reference sources?
spacediver is offline  
post #20 of 82 Old 10-25-2013, 02:53 PM
AVS Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 2,096
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 45 Post(s)
Liked: 72
Quote:
Originally Posted by spacediver View Post

I've just sent a request for further information, including pricing. Will report back if and when I hear from them.

If it turns out that the colorimetry specs are included, and are stable and accurate, then I don't understand why you couldn't use a set of these to test your instruments. Why on earth would you need a reference instrument, when you have reference sources?

It's a high luminance LED. Why would you expect it to be reference? Sure, it calls itself reference, but the only number it lists in the description is for luminance. It could very well be meant for use in testing light sensors in manufacturing, not for evaluation of video test equipment, where accurate colorimetry is also a requirement. Without detailed specs for xyY, and a manufacturer's guarantee that the item will meet those specs within so many percent, it's a crapshoot.

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931
Rolls-Royce is online now  
post #21 of 82 Old 10-25-2013, 03:03 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Quote:
Originally Posted by Rolls-Royce View Post

It's a high luminance LED. Why would you expect it to be reference? Sure, it calls itself reference, but the only number it lists in the description is for luminance. It could very well be meant for use in testing light sensors in manufacturing, not for evaluation of video test equipment, where accurate colorimetry is also a requirement. Without detailed specs for xyY, and a manufacturer's guarantee that the item will meet those specs within so many percent, it's a crapshoot.


this is exactly why I said:

Quote:
If it turns out that the colorimetry specs are included, and are stable and accurate...
spacediver is offline  
post #22 of 82 Old 10-25-2013, 03:10 PM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,518
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 169 Post(s)
Liked: 562
Quote:
Originally Posted by Rolls-Royce View Post

Quote:
Originally Posted by spacediver View Post

I've just sent a request for further information, including pricing. Will report back if and when I hear from them.

If it turns out that the colorimetry specs are included, and are stable and accurate, then I don't understand why you couldn't use a set of these to test your instruments. Why on earth would you need a reference instrument, when you have reference sources?

It's a high luminance LED. Why would you expect it to be reference? Sure, it calls itself reference, but the only number it lists in the description is for luminance. It could very well be meant for use in testing light sensors in manufacturing, not for evaluation of video test equipment, where accurate colorimetry is also a requirement. Without detailed specs for xyY, and a manufacturer's guarantee that the item will meet those specs within so many percent, it's a crapshoot.

+1

One of the lamps that JETI is using is a Mikropack HgAr lamp CAL-2000 (Calibration Light Source) with the following wavelengths:

365.02 nm; 404.66 nm; 435.84 nm; 546.07 nm; 696.54 nm;
706.72 nm; 727.29 nm; 738.40 nm; 763.51 nm; 772.40 nm;
794.82 nm; 811.53 nm; 826.45 nm; 852.14 nm; 912.30 nm.

comparing the results by using a Konica Minolta CS-2000 as a reference.

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is online now  
post #23 of 82 Old 10-25-2013, 09:12 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Just heard back from them - the light sources do indeed come with SPD, chromaticity, etc. and the calibration of these units can be traced to NIST. I'm not going to share the pricing information until I've been given permission to do so, but these lights should serve as an appopriate reference source to calibrate equipment.

After all, how do you think high end lab grade spectroradiometers are calibrated in the first place?

So, assuming one could afford these units, and assuming that their SPDs remain consistent over the 100 hours of operation they claim, one could conceivably do their own calibration check without the need of a reference instrument.

I have yet to hear from anyone in this thread why the above statement is not true.
spacediver is offline  
post #24 of 82 Old 10-25-2013, 09:38 PM
AVS Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 2,096
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 45 Post(s)
Liked: 72
Quote:
Originally Posted by spacediver View Post

Just heard back from them - the light sources do indeed come with SPD, chromaticity, etc. and the calibration of these units can be traced to NIST. I'm not going to share the pricing information until I've been given permission to do so, but these lights should serve as an appopriate reference source to calibrate equipment.

After all, how do you think high end lab grade spectroradiometers are calibrated in the first place?

So, assuming one could afford these units, and assuming that their SPDs remain consistent over the 100 hours of operation they claim, one could conceivably do their own calibration check without the need of a reference instrument.

I have yet to hear from anyone in this thread why the above statement is not true.

High grade lab spectros are usually calibrated using an integrating sphere.

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931
Rolls-Royce is online now  
post #25 of 82 Old 10-25-2013, 10:28 PM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Yes, if you have access to a full sized integrating sphere, then that's great. From what I'm told by the company, these LEDs are actually used to calibrate integrating spheres. I'm guessing this is this is for adjusting the port area sizes for measurements of luminous flux?

I suppose there's a chicken and egg situation here - what is the ultimate colorimetric reference - is it the source, or the spectro...

my point is - if these leds are what they claim to be, then I see no reason why they couldn't serve as a calibration reference for spectro/colorimeter.

I actually found a DIY integrating sphere that looked pretty cool, though I wouldn't trust it over one that had proper baffles and a high end diffusing surface.

Suppose you did trust the reference source, and the SPD is invariant over measuring angles. Would a pitch black room suffice?
spacediver is offline  
post #26 of 82 Old 10-26-2013, 12:33 AM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
btw, here is the original reply I got from them:

Thank you very much your inquiry; IPR-DES4 Reference Standard LEDs come with NIST tracable calibration for spectral power distribution, chromaticity, total power and intensity. They are used for calibrating spectroradiometers, integrating spheres, colorimeters. You can have them in any color.

Price of each IPR-DES4 HB LED is $4500, including calibration, power controller, leads. Delivery time after ordering 4-6 weeks.
spacediver is offline  
post #27 of 82 Old 10-26-2013, 12:55 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,518
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 169 Post(s)
Liked: 562
Quote:
Originally Posted by spacediver View Post

btw, here is the original reply I got from them:

Thank you very much your inquiry; IPR-DES4 Reference Standard LEDs come with NIST tracable calibration for spectral power distribution, chromaticity, total power and intensity. They are used for calibrating spectroradiometers, integrating spheres, colorimeters. You can have them in any color.

Price of each IPR-DES4 HB LED is $4500, including calibration, power controller, leads. Delivery time after ordering 4-6 weeks.

Nice Info, but it's beyond to that the user asked, all your reply's here are useless for him, you provide zero help to the user question with the used i1PRO.

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is online now  
post #28 of 82 Old 10-26-2013, 01:06 AM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Quote:
Originally Posted by ConnecTEDDD View Post

Nice Info, but it's beyond to that the user asked, all your reply's here are useless for him, you provide zero help to the user question with the used i1PRO.

I went out of my way to provide an idea that hadn't been suggested, actually contact the company and ask for a quote (which may be of value to others reading this thread), engage in an interesting theoretical discussion that might be of interest to others, and you think it appropriate to criticize me because it didn't end up helping him??

I could see how this would be a valid criticism had I charged him for any help, or had I criticized other valid suggestions.

I don't understand the purpose of your post.
spacediver is offline  
post #29 of 82 Old 10-26-2013, 01:21 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,518
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 169 Post(s)
Liked: 562
Quote:
Originally Posted by spacediver View Post

Quote:
Originally Posted by ConnecTEDDD View Post

Nice Info, but it's beyond to that the user asked, all your reply's here are useless for him, you provide zero help to the user question with the used i1PRO.

I went out of my way to provide an idea that hadn't been suggested, actually contact the company and ask for a quote (which may be of value to others reading this thread), engage in an interesting theoretical discussion that might be of interest to others, and you think it appropriate to criticize me because it didn't end up helping him??

I could see how this would be a valid criticism had I charged him for any help, or had I criticized other valid suggestions.

I don't understand the purpose of your post.

The problem with your currect posts here is that you suggested a solution that you was sure about it's successfull results, but the same time you had no actual information about you the cost and the way that someone can try this at home.

If you like you can open a new thread to discus that lamp solution you found for us to stop sending our meters to X-Rite's Labs anymore, to find out a cheaper lamp etc.

But you can write whatever you want, it's a public forum. wink.gif I'm not here to judge you.

Personally I refer to inform with things i know, not with imaginary things what might work.

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is online now  
post #30 of 82 Old 10-26-2013, 01:30 AM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 948
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 122 Post(s)
Liked: 77
Goodnight Ted, I have no wish to further engage you in this discussion.
spacediver is offline  