post #91 of 137 Old 01-30-2014, 07:34 PM
Advanced Member
 
gwgill's Avatar
 
Join Date: Jan 2013
Location: Melbourne, Australia
Posts: 594
Mentioned: 3 Post(s)
Tagged: 0 Thread(s)
Quoted: 51 Post(s)
Liked: 75
Quote:
Originally Posted by Jeroen1000 View Post

If there is a problem, just copying a lot of files to a portable hard drive will reveal it. There are simple tools available to hash them and see whether they've arrived bit perfect.

This issue has nothing to do with bits getting across the cable, it is about the quality of the power that is supplying the analog electronics around the sensor, the A./D converter and the reference light source.
ConnecTEDDD likes this.
gwgill is offline  
Sponsored Links
Advertisement
 
post #92 of 137 Old 02-03-2014, 05:22 PM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
JETI has updaTED their website with a new Application Note Document:

There some specific settings that may prevent communication problem, interesting...

Summary of Measures in Case of USB Communication Problems (Application Note 23)

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #93 of 137 Old 02-04-2014, 01:17 AM - Thread Starter
AVS Special Member
 
sillysally's Avatar
 
Join Date: Jun 2006
Location: Chicago
Posts: 3,714
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 73 Post(s)
Liked: 262
Quote:
Originally Posted by ConnecTEDDD View Post

JETI has updaTED their website with a new Application Note Document:

There some specific settings that may prevent communication problem, interesting...

Summary of Measures in Case of USB Communication Problems (Application Note 23)

Interesting.

Ted do you know when Jeti is talking about a external USB power supply, are they talking about a power hub. ?

I was playing with the 1211 today on my VT60. It seems that setting the refresh rate at 60Hz rather than 96Hz along with setting my Lumagen to out put video RGB BT 709 rather than 4:2:2 BT 709 stabilizes the Red readings. Yes I am using the auto sync for the 1211.

I will have to play some more to determine what works best with the VT60 and the 1211. I will also use Jeti advice on the settings for USB in my PC tomorrow.

Right now I am running a 21^3 profile. I profiled my K10-A with the 1211, setup my 100% grayscale setting in my VT60. The interesting thing for me was that I have always felt Red was off (to high) when setting my RGB in my VT60. Before I ran the 21^3 profile I set my 100% setting in my VT60, and yes it seems that when using the 1211 stand alone or the K10 profiled to the 1211 Red was being set to high when using my I1Pro 2 to profile my K10.

ss
sillysally is offline  
post #94 of 137 Old 02-04-2014, 01:42 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by sillysally View Post

Quote:
Originally Posted by ConnecTEDDD View Post

JETI has updaTED their website with a new Application Note Document:

There some specific settings that may prevent communication problem, interesting...

Summary of Measures in Case of USB Communication Problems (Application Note 23)

Interesting.

Ted do you know when Jeti is talking about a external USB power supply, are they talking about a power hub. ?

I was playing with the 1211 today on my VT60. It seems that setting the refresh rate at 60Hz rather than 96Hz along with setting my Lumagen to out put video RGB BT 709 rather than 4:2:2 BT 709 stabilizes the Red readings. Yes I am using the auto sync for the 1211.

I will have to play some more to determine what works best with the VT60 and the 1211. I will also use Jeti advice on the settings for USB in my PC tomorrow.

Right now I am running a 21^3 profile. I profiled my K10-A with the 1211, setup my 100% grayscale setting in my VT60. The interesting thing for me was that I have always felt Red was off (to high) when setting my RGB in my VT60. Before I ran the 21^3 profile I set my 100% setting in my VT60, and yes it seems that when using the 1211 stand alone or the K10 profiled to the 1211 Red was being set to high when using my I1Pro 2 to profile my K10.

ss


SS my friend, aks JETI to tell you if they are suggesting this linear power supply or if they have tesTED it:

AQVOX USB Low-Noise 5V Isolated Linear Power Supply

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #95 of 137 Old 02-04-2014, 03:55 AM
AVS Addicted Member
 
Join Date: Sep 2002
Location: Wetumpka, AL
Posts: 15,347
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 104 Post(s)
Liked: 190
Quote:
Originally Posted by sillysally View Post

....snip...
The interesting thing for me was that I have always felt Red was off (to high) when setting my RGB in my VT60. Before I ran the 21^3 profile I set my 100% setting in my VT60, and yes it seems that when using the 1211 stand alone or the K10 profiled to the 1211 Red was being set to high when using my I1Pro 2 to profile my K10.

ss

How does that compare to the 1201s you had?

Samsung 64F8500, Panasonic 65VT50, Oppo 95, Tivo Roamio for OTA, Dish VIP722, Marantz AV8801 preamp, Rotel Amps, Atlantic Tech 8200 speakers, Seaton Submersive HP, Calman 5, Chromapure, Accupel DVG-5000, i1Display3pro, i1pro2, eecolor colorbox.
JimP is online now  
post #96 of 137 Old 02-04-2014, 07:22 AM - Thread Starter
AVS Special Member
 
sillysally's Avatar
 
Join Date: Jun 2006
Location: Chicago
Posts: 3,714
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 73 Post(s)
Liked: 262
Quote:
Originally Posted by JimP View Post

How does that compare to the 1201s you had?

Its not that the 1201 is a bad mete,r probably a very good meter. It can't sync with my VT60 or other Plasma's from what is said.

I need to play more with the 1211, but yes it seems to sync with my VT60.
As I thought Red was off the biggest, then Green and Magenta using my I1Pro 2 to profile the K10-A.

When running CM's QC every thing that has to do with the CMS (Color points, Lum, Sat) is tighter along with fine tuning the grayscale.
The little I have viewed a Blu Ray disc, there hasn't been anything that looks slightly off.

I can see when I am are setting my black level at 0% and 5% you still need the K10. Like the CS 650,850 and 2000 they are very slow reading low light, so the K10 is a very nice plus.
There is no way I could do a 21^3 profile/LUT using just the 1211 or any spectro only. So at the very least you would need a D3 to go along with the 1211.

I hope this was not just a fluke, so I will be able to tell more in a week or so.

ss
sillysally is offline  
post #97 of 137 Old 02-04-2014, 09:02 AM
AVS Addicted Member
 
Join Date: Sep 2002
Location: Wetumpka, AL
Posts: 15,347
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 104 Post(s)
Liked: 190
SS

It sounds like the i1d3 is faster than the 1211 at low light. Is that your take on it?

Samsung 64F8500, Panasonic 65VT50, Oppo 95, Tivo Roamio for OTA, Dish VIP722, Marantz AV8801 preamp, Rotel Amps, Atlantic Tech 8200 speakers, Seaton Submersive HP, Calman 5, Chromapure, Accupel DVG-5000, i1Display3pro, i1pro2, eecolor colorbox.
JimP is online now  
post #98 of 137 Old 02-04-2014, 09:02 AM
Member
 
DrFaxe's Avatar
 
Join Date: Nov 2012
Posts: 192
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 1 Post(s)
Liked: 76
Quote:
Originally Posted by sillysally View Post


snip

As I thought Red was off the biggest, then Green and Magenta using my I1Pro 2 to profile the K10-A.

When running CM's QC every thing that has to do with the CMS (Color points, Lum, Sat) is tighter along with fine tuning the grayscale.
The little I have viewed a Blu Ray disc, there hasn't been anything that looks slightly off.

snip

ss


Hi SillySally,

I'm getting a bit nervous when reading Your post about the i1pro2, because I just ordered one to calibrate my VT60 along with the I1D3.
Could You say anything about how much the i1Pro2 is off compared to the Jeti?

thanks!!
ConnecTEDDD likes this.
DrFaxe is offline  
post #99 of 137 Old 02-04-2014, 09:43 AM - Thread Starter
AVS Special Member
 
sillysally's Avatar
 
Join Date: Jun 2006
Location: Chicago
Posts: 3,714
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 73 Post(s)
Liked: 262
Quote:
Originally Posted by JimP View Post

SS

It sounds like the i1d3 is faster than the 1211 at low light. Is that your take on it?

My sold C6 (same as D3) was faster on low light.

Quote:
Originally Posted by DrFaxe View Post

Hi SillySally,

I'm getting a bit nervous when reading Your post about the i1pro2, because I just ordered one to calibrate my VT60 along with the I1D3.
Could You say anything about how much the i1Pro2 is off compared to the Jeti?

thanks!!

Don't worry the I1pro 2 is a good spectro, just not reference. There are a lot of "pros" that use the I1Pro,

The picture was done by Klein, using there reference CS2000 spectro and my I1Pro 2


ss
ConnecTEDDD likes this.
sillysally is offline  
post #100 of 137 Old 02-04-2014, 10:31 AM
Member
 
DrFaxe's Avatar
 
Join Date: Nov 2012
Posts: 192
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 1 Post(s)
Liked: 76
Thanks for the chart.
Do You know what was the light source, TV? It would be interesting to see this with a VT60.
As to the new red phosphor Panasonic is using I would expect the colors (red) will show different (smaller) bandwidth according to the VT50.
Couldn't this cause the issue with the i1pro2 on red?
Do You remember if the colors were off in the same way on the VT50 calibrated with i1pro2?

Can You please provide a chart like above using Your Jeti and i1po2 with the VT60?
Really would appreciate that.
ConnecTEDDD likes this.
DrFaxe is offline  
post #101 of 137 Old 02-04-2014, 10:32 AM
Advanced Member
 
realzven's Avatar
 
Join Date: Mar 2010
Posts: 519
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 84
the i1 pro seems to reads Y lower than the cs2000
ConnecTEDDD likes this.
realzven is offline  
post #102 of 137 Old 02-04-2014, 05:13 PM
AVS Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 6,399
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 40 Post(s)
Liked: 211
Quote:
Originally Posted by JimP View Post

SS

It sounds like the i1d3 is faster than the 1211 at low light. Is that your take on it?
Virtually any colorimeter will be more sensitive than any spectroradiometer. Tristimulus colorimeters have much larger photo-sensitive elements.
ConnecTEDDD likes this.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #103 of 137 Old 02-04-2014, 06:04 PM
AVS Special Member
 
zoyd's Avatar
 
Join Date: Sep 2006
Location: Planet Dog
Posts: 4,451
Mentioned: 10 Post(s)
Tagged: 0 Thread(s)
Quoted: 91 Post(s)
Liked: 314
zoyd is online now  
post #104 of 137 Old 02-04-2014, 07:19 PM
AVS Addicted Member
 
Join Date: Sep 2002
Location: Wetumpka, AL
Posts: 15,347
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 104 Post(s)
Liked: 190
Quote:
Originally Posted by zoyd View Post

Here is another data point comparing the i1pro2 to JETI-1211



zoyd....was this off of a plasma display?

Samsung 64F8500, Panasonic 65VT50, Oppo 95, Tivo Roamio for OTA, Dish VIP722, Marantz AV8801 preamp, Rotel Amps, Atlantic Tech 8200 speakers, Seaton Submersive HP, Calman 5, Chromapure, Accupel DVG-5000, i1Display3pro, i1pro2, eecolor colorbox.
JimP is online now  
post #105 of 137 Old 02-05-2014, 02:10 AM
Member
 
DrFaxe's Avatar
 
Join Date: Nov 2012
Posts: 192
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 1 Post(s)
Liked: 76
Quote:
Originally Posted by zoyd View Post

Here is another data point comparing the i1pro2 to JETI-1211

thanks for the data Zoyd. Same question as JimP here, what was the light source?
DrFaxe is offline  
post #106 of 137 Old 02-05-2014, 03:45 AM
AVS Special Member
 
zoyd's Avatar
 
Join Date: Sep 2006
Location: Planet Dog
Posts: 4,451
Mentioned: 10 Post(s)
Tagged: 0 Thread(s)
Quoted: 91 Post(s)
Liked: 314
zoyd is online now  
post #107 of 137 Old 02-07-2014, 08:14 AM
Member
 
Kukulcan's Avatar
 
Join Date: Sep 2012
Location: Italy
Posts: 117
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 20
Quote:
Originally Posted by sillysally View Post


Don't worry the I1pro 2 is a good spectro, just not reference.

a reference spectro (with higher resolution) becomes more important for measuring the spectro of very wide gamut displays, because the wider the gamut, the narrower must be the RGB peaks, and so the higher resolution must have the spectrometer in order to measure accurately those peaks.

I think that if you compare the iPro vs CS2000 when measuring a Vango vpr (RGB LED), you'd see larger measurement differences.
Kukulcan is offline  
post #108 of 137 Old 02-07-2014, 12:31 PM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Here is an example of the Spectral Response Resolution of i1PRO vs. JETI 1211 measuring a Pioneer KURO.



Here is an another example of i1PRO vs. JETI 1211 on a Panasonic Plasma... +2.0dE difference between meter readings on blue is easily visible

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #109 of 137 Old 02-07-2014, 01:16 PM
Senior Member
 
CalWldLif's Avatar
 
Join Date: Feb 2011
Location: South Bay'Los Angeles California.
Posts: 462
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 43 Post(s)
Liked: 26
Quote:
Originally Posted by ConnecTEDDD View Post

Here is an example of the Spectral Response Resolution of i1PRO vs. JETI 1211 measuring a Pioneer KURO.



Here is an another example of i1PRO vs. JETI 1211 on a Panasonic Plasma... +2.0dE difference between meter readings on blue is easily visible
interesting.
In my limited supply of meters, blue is the one color that changes the most from meter to meter.

Loving D65
CalWldLif is offline  
post #110 of 137 Old 02-07-2014, 03:27 PM
AVS Special Member
 
zoyd's Avatar
 
Join Date: Sep 2006
Location: Planet Dog
Posts: 4,451
Mentioned: 10 Post(s)
Tagged: 0 Thread(s)
Quoted: 91 Post(s)
Liked: 314
Quote:
Originally Posted by ConnecTEDDD View Post

Here is an example of the Spectral Response Resolution of i1PRO vs. JETI 1211 measuring a Pioneer KURO.


Here is an another example of i1PRO vs. JETI 1211 on a Panasonic Plasma... +2.0dE difference between meter readings on blue is easily visible

Hi Ted,

Please explain why anyone would care about a 2 dE difference between the two readings and address the following:

1. 2 dE may be detectable in side by side still images but in moving images it will not be noticeable.
2. The measurement is at fully saturated blue which will hardly ever occur in typical material.
zoyd is online now  
post #111 of 137 Old 02-07-2014, 11:45 PM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by zoyd View Post

Quote:
Originally Posted by ConnecTEDDD View Post

Here is an example of the Spectral Response Resolution of i1PRO vs. JETI 1211 measuring a Pioneer KURO.


Here is an another example of i1PRO vs. JETI 1211 on a Panasonic Plasma... +2.0dE difference between meter readings on blue is easily visible

Hi Ted,

Please explain why anyone would care about a 2 dE difference between the two readings and address the following:

1. 2 dE may be detectable in side by side still images but in moving images it will not be noticeable.
2. The measurement is at fully saturated blue which will hardly ever occur in typical material.

Hello Zoyd,

Any Meter Profiling Measurement Patch Sequence (Spectro -> Colorimeter) is using RGB 100% Saturated Color Patterns, so if you have a difference of 2dE there, this is translaTED to some milion shades blue that will be of by off by 2.0dE minimum.

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #112 of 137 Old 02-08-2014, 01:07 AM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 796
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 50 Post(s)
Liked: 66
Quote:
Originally Posted by ConnecTEDDD View Post

translaTED

really?
spacediver is offline  
post #113 of 137 Old 02-08-2014, 01:19 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by spacediver View Post

Quote:
Originally Posted by ConnecTEDDD View Post

translaTED

really?

What do you mean?

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #114 of 137 Old 02-08-2014, 01:23 AM
Advanced Member
 
spacediver's Avatar
 
Join Date: May 2013
Location: Toronto
Posts: 796
Mentioned: 1 Post(s)
Tagged: 0 Thread(s)
Quoted: 50 Post(s)
Liked: 66
You capitalize the letters "ted" whenever they occur in your posts. It just seems a bit... silly?
spacediver is offline  
post #115 of 137 Old 02-08-2014, 01:27 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by spacediver View Post

You capitalize the letters "ted" whenever they occur in your posts. It just seems a bit... silly?

Your post is out of topic here, I'm not forcing anyone to read my posts. You can ignore them if it's dificult for you to read them. Thanks

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #116 of 137 Old 02-08-2014, 03:15 AM
AVS Addicted Member
 
Join Date: Sep 2002
Location: Wetumpka, AL
Posts: 15,347
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 104 Post(s)
Liked: 190
Ted,

Your point about being off 2dEs throughout a cube calibration when your profile at 100% would seem to not really matter when taken in isolation but I'd wager that other things may be going on that would add addition dEs such as plasma panel fatigue, meter drift, etc. when running a calibration that takes an hour and a half up to four hours with high end color meters.(too long a sentence smile.gif ) So it does make sense to get it as accurate as possible as other things will creep into the measurements.

With that said, traditional calibrations wouldn't have as much need since you have only a few measurement point and can pause your calibration and run screen wipes or full screen white patterns to clean up IR.

Since large cube calibrations do take a bit of time to perform, it wouldn't be a bad idea to add the feature where cube calibration periodically pauses while patterns display that would clean up IR and return the display to normal operating temperatures. Although this is mostly a plasma phenomenon, who really knows if OLED or any future tech wouldn't also benefit from this.

Nothing personal, but I think spacediver is right. Brother, it's time to leave it in the past.
Plutotype likes this.

Samsung 64F8500, Panasonic 65VT50, Oppo 95, Tivo Roamio for OTA, Dish VIP722, Marantz AV8801 preamp, Rotel Amps, Atlantic Tech 8200 speakers, Seaton Submersive HP, Calman 5, Chromapure, Accupel DVG-5000, i1Display3pro, i1pro2, eecolor colorbox.
JimP is online now  
post #117 of 137 Old 02-08-2014, 11:07 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by JimP View Post

Ted,

Your point about being off 2dEs throughout a cube calibration when your profile at 100% would seem to not really matter when taken in isolation but I'd wager that other things may be going on that would add addition dEs such as plasma panel fatigue, meter drift, etc. when running a calibration that takes an hour and a half up to four hours with high end color meters.(too long a sentence smile.gif ) So it does make sense to get it as accurate as possible as other things will creep into the measurements.

With that said, traditional calibrations wouldn't have as much need since you have only a few measurement point and can pause your calibration and run screen wipes or full screen white patterns to clean up IR.

Since large cube calibrations do take a bit of time to perform, it wouldn't be a bad idea to add the feature where cube calibration periodically pauses while patterns display that would clean up IR and return the display to normal operating temperatures. Although this is mostly a plasma phenomenon, who really knows if OLED or any future tech wouldn't also benefit from this.

Nothing personal, but I think spacediver is right. Brother, it's time to leave it in the past.

Hello, If we include the fact that:

1) There is no standard in Gamma Value

2) You don't know if the displays are used for Blu-Mastering was accurate or how tight was calibrated.

3) What meters / software they used , are you using the same at home?

4) What was the enviroment light conditions during the mastering, do you have the same at home ?

5) What display types they used to verify the content.

6) What peak light output they used, what window patterns they used, did you used the same size or peak during your calibration?

7) What driftings they had to their monitors (since it was used for a lot of times continiously), and your own display drifting.

8) The fact that all displays/projectors are drifitng during the working time, while you see a movie 2-3 hours is operating.

9) Add the off angle problems or LED's, are you viewing from dead center?

10) Add the bad display screen uniformity of the consumer displays.

...after all the above seems like there is no need to try to get the best from your display.... put the movie mode , no calibrating, and all good...

But No, this is about image fidelity, using better tools and software you are getting closer.

Anyone with LightSpace who want to see the difference of 2dE to his setup, he can do the following:

Using a saved profiling file, generate 2 correction LUTs, one with REC709 targets and another one with modified REC709 targets, for example change one xy color cordinates to have 2.0 dE, and after load these LUT's to Lumagen or eeColor, swap the memories and see the difference.

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #118 of 137 Old 02-08-2014, 11:12 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by JimP View Post

Ted,

Your point about being off 2dEs throughout a cube calibration when your profile at 100% would seem to not really matter.

Do you believe the is no visible difference if you use i1PRO or JETI 1211? since the dE difference numbers are low?

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
post #119 of 137 Old 02-08-2014, 05:11 PM
Member
 
plasma_fan's Avatar
 
Join Date: Jan 2012
Posts: 53
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 14
Maybe not the point but;

This dE of 2 shown compares a single i1pro to a single jeti 1211 on a single display.
Were the same two meters that were compared on the Kuro used, if so what is different about the Panasonic?
Were readings done back to back?
Is it just assumed that the 1211 is the reference? was it verified?
Is it possible that the 1211 was reading 1 dE low and the i1pro reading 1 dE high?
When was the i1pro last certified?
Can (has) a similar result been duplicated with a different pair of meters on a different panasonic of the same model?
Specifically what Panasonic plasma model?
And as zoyd already pointed out will the same difference between meters be present at a lesser saturation point?

No doubt in my mind what meter i think is better and which meter i would blindly trust more, but there are more questions than answers that come from this flashing chart.
plasma_fan is offline  
post #120 of 137 Old 02-09-2014, 06:37 AM
AVS Special Member
 
ConnecTEDDD's Avatar
 
Join Date: Sep 2010
Location: Athens, Greece
Posts: 2,270
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 82 Post(s)
Liked: 462
Quote:
Originally Posted by plasma_fan View Post

Maybe not the point but;

This dE of 2 shown compares a single i1pro to a single jeti 1211 on a single display.

Were the same two meters that were compared on the Kuro used, if so what is different about the Panasonic?

Yes, same meters, some time, same place

Were readings done back to back?

Yes, BTW

Is it just assumed that the 1211 is the reference? was it verified?

Yes, JETI 1211 (11.000$) is reference for sure, it was brand new.

Is it possible that the 1211 was reading 1 dE low and the i1pro reading 1 dE high?
When was the i1pro last certified?

i1PRO was inside it's 1 year certification period.

Can (has) a similar result been duplicated with a different pair of meters on a different panasonic of the same model?

I don't understand why you need this ? It's measurements, and JETI is 11 times more expensive meter than i1PRO, do you expect to have the same performance?

Specifically what Panasonic plasma model?

Panasonic ST60

And as zoyd already pointed out will the same difference between meters be present at a lesser saturation point?

Ask Zoyd to send you data for this, next month I will have JETI 1211 here with me and I will compare it with my i1PRO1 + 2

No doubt in my mind what meter i think is better and which meter i would blindly trust more, but there are more questions than answers that come from this flashing chart.

Hello, Check @ bold letters above is my reply.

Ted's LightSpace CMS Calibration Disk Free Version for Free Calibration Software: LightSpace DPS + CalMAN ColorChecker
S/W: LightSpace CMS, SpaceMan ICC, SpaceMatch DCM, CalMAN 5, CalMAN RGB, ChromaPure, CalPC, ControlCAL
Meters: JETI Specbos 1211, Klein K-10A, i1PRO2, i1PRO, SpectraCAL C6, i1D3, C5
ConnecTEDDD is offline  
Reply Display Calibration

User Tag List

Thread Tools
Show Printable Version Show Printable Version
Email this Page Email this Page


Forum Jump: 

Posting Rules  
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off