I thought it was time to post some more formal testing I have done of three popular calibration probes:
  • X-Rite i1Pro spectroradiometer
  • X-Right Display 2 colorimeter
  • DataColor Spyder2 colorimeter
I decided to use the i1Pro as the reference. It is the most expensive meter, the best design, and has the best reputation. I recently had an opportunity to compare the i1Pro against a much more expensive spectroradiometer. The results were very similar, and served to reinforce my judgment that the i1Pro offers excellent performance. It's only shortcomings are relatively poor low light level performance and the annoying need to take regular dark readings to maintain accuracy.

The results offered some surprises. In no particular order:
  • Colorimeters are NOT more subject to errors with chromaticity readings. In fact, the Display 2 in particular can often offer excellent performance for RGBYCM readings. I found larger errors with gray scale readings.
  • The biggest drawback of colorimeters may NOT be that they offer particularly inaccurate results with some types of displays. Based on my testing, their biggest weakness seems to be inter-probe variability. Before I began the test I purchased a new Spyder2 and Display 2. They both measured substantially differently from the meters they replaced. For example, I had previously found that the Display 2 had a problem with plasmas, but this probe seemed to perform quite well on the same plasma I had previously tested. It also seemed to perform a little worse with the DLP and CRT.
  • The Spyder2 performed substantially worse than the Display 2 in almost every test.
  • Colorimeters do NOT seem especially optimized for CRTs. Their accuracy on CRT seemed about average: better than with some digital displays and worse than with others.
As I have written before, the performance of the Spyder2 is not suitable for serious work, especially with the availability of the much superior and quite inexpensive Display 2.

Here are the detailed results from 6 displays.

CRT



Plasma



DLP



SXRD (LCoS)



LCD 1



LCD 2



As you can see, the worst results were obtained from the colorimeters for gray scale readings on an SXRD rear projector. The chromaticity readings were much better. The best performance was obtained on one of the LCDs. Interestingly, the second worst performance occurred on the other LCD, indicating some variable beyond display technology.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
Sponsored Links
Advertisement
 
post #2 of 315 Old 12-19-2007, 06:41 PM
Advanced Member
 
sperron's Avatar
 
Join Date: Dec 2006
Posts: 847
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
Thanks for doing this!

I wasn't very happy with the gray scale the Display 2 did on my SXRD, and I can see why now. I'll have to look into getting a eye-one pro. Any recommendations on getting one for a reasonable price?
sperron is offline  
post #3 of 315 Old 12-19-2007, 07:37 PM
AVS Special Member
 
greeno's Avatar
 
Join Date: Mar 2002
Location: Livermore, CA
Posts: 2,592
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 19
Hi Tom,
What software did you use to interface with the meters? Are the results different outputing the measurements in xyz instead of xyY? I'm asking really about the accuracy of internal calculations of the meters. Some have settings for extended precision arithmetic that could affect the results.

Best and thanks for the results.
Best,
jeff
greeno is offline  
post #4 of 315 Old 12-19-2007, 07:38 PM
AVS Special Member
 
Joelc's Avatar
 
Join Date: Apr 2001
Location: Toronto, Ontario, Canada
Posts: 2,067
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
Tom:

Great thread...the one issue that I have with your analysis is that you have used the i1Pro as the reference which creates a problems...that being that dE is measured against the i1Pro and does not really indicate which is truly more accurate..

As a result of your testing what equipment would you suggest one who has both a colorimeter and a sprectroradiometer for each of grayscale and gamut.

TIA.

Joel
Some people choose to have a pool, I choose to have a home theater!
Joelc is offline  
post #5 of 315 Old 12-19-2007, 09:51 PM - Thread Starter
AVS Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 6,353
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 14 Post(s)
Liked: 204
I ran the i1Pro with ColorFacts, Progressive Labs, and HCFR. The results did not vary more than 0.001.

Quote:
Originally Posted by greeno View Post

Hi Tom,
What software did you use to interface with the meters? Are the results different outputing the measurements in xyz instead of xyY? I'm asking really about the accuracy of internal calculations of the meters. Some have settings for extended precision arithmetic that could affect the results.

Best and thanks for the results.
Best,
jeff


Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #6 of 315 Old 12-19-2007, 09:53 PM - Thread Starter
AVS Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 6,353
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 14 Post(s)
Liked: 204
Right now the i1Pro is the closest to a reference that I have. The Display 2 works very well measuring gamut, but I'd stick with the i1Pro for grayscale.

Quote:
Originally Posted by Joelc View Post

Tom:

Great thread...the one issue that I have with your analysis is that you have used the i1Pro as the reference which creates a problems...that being that dE is measured against the i1Pro and does not really indicate which is truly more accurate..

As a result of your testing what equipment would you suggest one who has both a colorimeter and a sprectroradiometer for each of grayscale and gamut.

TIA.


Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #7 of 315 Old 12-20-2007, 03:45 AM
AVS Special Member
 
Joelc's Avatar
 
Join Date: Apr 2001
Location: Toronto, Ontario, Canada
Posts: 2,067
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
Quote:
Originally Posted by TomHuffman View Post

Right now the i1Pro is the closest to a reference that I have. The Display 2 works very well measuring gamut, but I'd stick with the i1Pro for grayscale.

Thanks for the feedback...

Joel
Some people choose to have a pool, I choose to have a home theater!
Joelc is offline  
post #8 of 315 Old 12-20-2007, 07:47 AM
AVS Special Member
 
alluringreality's Avatar
 
Join Date: Jul 2006
Posts: 3,113
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 3 Post(s)
Liked: 30
Quote:
Originally Posted by sperron View Post

I wasn't very happy with the gray scale the Display 2 did on my SXRD, and I can see why now. I'll have to look into getting a eye-one pro.

Tom doesn't mention how exactly he made the measurements or the model, but one thing about both SXRDs I've owned is that grayscale measurements will vary if you move the meter much at all. So for a comparison on SXRD you would have to be very careful in using the same meter position to compare grayscale so that you could give some idea of how close the meters were positioned. Certainly I think marking the screen would be absolutely necessary in making a grayscale comparison on SXRD. The primaries don't seem to vary much regardless of meter position, so it doesn't matter there.

I'm sure my Display LT has some inaccuracy, but without some idea of how much the meters might have been set in different locations, then I can't really test how much variation an individual meter can show by that movement. On all the talk about how a colorimeter is inaccurate for SXRD, no one has listed how much their measuring location might have varied or brings up the non-uniformity issue. The first addresses just how much variation each meter might show by the difference in positioning and the second brings into questioning how much a more accurate device really matters when the screen isn't uniform in the first place.
alluringreality is offline  
post #9 of 315 Old 12-20-2007, 12:09 PM
Advanced Member
 
-Hitman-'s Avatar
 
Join Date: Nov 2004
Location: Manchester, UK
Posts: 626
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 20 Post(s)
Liked: 50
Quote:
Originally Posted by TomHuffman View Post

I ran the i1Pro with ColorFacts, Progressive Labs, and HCFR. The results did not vary more than 0.001.

Hi Tom,

Slightly off topic but..

I see you've mentioned that you use colourfacts, is it the version that contains the C3 interface/protocol for Pioneer Plasmas to enable day/night C3 modes?

Reason i ask is that Georges in the HCFR forum is after information on what the protocol is, so that he can possibly add it to the excellent HCFR.

Can you help?


TIA.
-Hitman- is online now  
post #10 of 315 Old 12-20-2007, 03:03 PM
Advanced Member
 
sperron's Avatar
 
Join Date: Dec 2006
Posts: 847
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
I can clearly see that my gray scale with the display 2 is definitely not on track even though the meter is reporting .313 .329 . Of course it could be that my D2 has drifted and just needs to be replaced. I'd rather replace it with an i1 pro, assuming I can get one for a reasonable price.

I can see where you'd want to ask Tom about his methodology though. I know that moving my D2 as little as a quarter inch can cause the meter to get different readings on my SXRD. I've rigged mine up to a tripod, but even so I can't directly compare one calibration to another since I can't guarantee the meter is positioned exactly the same.
sperron is offline  
post #11 of 315 Old 12-20-2007, 03:47 PM
Advanced Member
 
sync's Avatar
 
Join Date: Sep 2000
Posts: 827
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 12
Tom,

Is the Display 2 you used the Eye-One model?

Also, my understanding is that these devices are designed for monitors. Will they work for LCD TVs that have much larger pixels than a monitor? Doesn't the colorimeter fit over colored squares that are displayed on the screen? It seems that on a large tv some of the squares would be displayed beyond the perimeter of the sensor.
sync is offline  
post #12 of 315 Old 12-21-2007, 10:41 AM
Member
 
alwilli's Avatar
 
Join Date: Jun 2003
Location: Huntsville, AL
Posts: 111
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 11
Tom,

Your luminance (Y) values for the DLP runs seem quite low. What units are they?

Alvin
alwilli is offline  
post #13 of 315 Old 12-22-2007, 02:46 PM
Advanced Member
 
sperron's Avatar
 
Join Date: Dec 2006
Posts: 847
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
For anyone interested in getting a good deal on an Eye-One pro, I found an authorized reseller site that has it in the "Photo LT" kit for $732 shipped (USA and Puerto Rico only). You then can print out a rebate form and get either $200 back from x-rite or a free beamer upgrade kit. This offer only applies to "Photo LT" kits purchased by 12-31-07. So basically a brand new Eye-One pro for $532.

https://www.digitalgraphicsresources...OD&ProdID=2091
sperron is offline  
post #14 of 315 Old 12-22-2007, 03:04 PM
AVS Special Member
 
zoyd's Avatar
 
Join Date: Sep 2006
Location: Planet Dog
Posts: 4,427
Mentioned: 6 Post(s)
Tagged: 0 Thread(s)
Quoted: 75 Post(s)
Liked: 299
Quote:
Originally Posted by TomHuffman View Post

Based on my testing, their biggest weakness seems to be inter-probe variability. Before I began the test I purchased a new Spyder2 and Display 2. They both measured substantially differently from the meters they replaced. For example, I had previously found that the Display 2 had a problem with plasmas, but this probe seemed to perform quite well on the same plasma I had previously tested. It also seemed to perform a little worse with the DLP and CRT.

Tom, very nice set of measurements. Were the D2 and S2 probe variations (new vs. old) inside or outside of their factory uncertainties? (S2: +/- 0.002(LCD) +/- 0.0035(CRT) and D2:+/- 0.004 in x,y neglecting Y)
zoyd is online now  
post #15 of 315 Old 12-22-2007, 06:20 PM - Thread Starter
AVS Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 6,353
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 14 Post(s)
Liked: 204
Zoyd:

I'm on vacation now. I'll respond to this and other questions when I return home after the holidays.

Quote:
Originally Posted by zoyd View Post

Tom, very nice set of measurements. Were the D2 and S2 probe variations (new vs. old) inside or outside of their factory uncertainties? (~ +/- 0.002 in x,y neglecting Y)


Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #16 of 315 Old 12-22-2007, 06:22 PM - Thread Starter
AVS Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 6,353
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 14 Post(s)
Liked: 204
It was a front projector. Measurements taken off the screen.

Quote:
Originally Posted by alwilli View Post

Tom,

Your luminance (Y) values for the DLP runs seem quite low. What units are they?

Alvin


Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #17 of 315 Old 12-22-2007, 09:55 PM
AVS Special Member
 
jdbimmer's Avatar
 
Join Date: Jan 2007
Posts: 1,003
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 14
Great test Tom. Too bad you didn't have a DTP94 on hand too. I have 2 here and an i1Pro. Although one DTP94 is much older than the other, some quick measurements showed them to return very similar results - I'll try to posts some numbers. Could be a fluke, but maybe this discontinued probe might actually have been the most consistent probe to probe and less vulnerable to variation over time.

Questions:
1. Were you able to to set the D2 to CRT mode for the CRT and Plasma measurements?
2. Did one of the LCD's have a so-called "wide gamut" light source?
jdbimmer is offline  
post #18 of 315 Old 12-23-2007, 07:30 AM
Member
 
hrc4u's Avatar
 
Join Date: Oct 2004
Location: sweden
Posts: 159
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 12
Tom.
can you get your hands on the new Spyder3 sensor and compare it against your other meters?..
/H
hrc4u is offline  
post #19 of 315 Old 12-23-2007, 08:16 AM
Advanced Member
 
anbjornk's Avatar
 
Join Date: Jun 2002
Location: Stavanger, Norway
Posts: 563
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
Quote:
Great test Tom. Too bad you didn't have a DTP94 on hand too.

I believe the DTP94 is basically the same as the Display2... (Correct me if I'm wrong)
anbjornk is offline  
post #20 of 315 Old 12-23-2007, 08:22 AM
AVS Special Member
 
richlo's Avatar
 
Join Date: Aug 2004
Posts: 1,463
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 11
Quote:
Originally Posted by anbjornk View Post

I believe the DTP94 is basically the same as the Display2... (Correct me if I'm wrong)


D2 is the successor but different than the DTP94,

Rich L

Go Chicago Bears

richlo is offline  
post #21 of 315 Old 12-23-2007, 08:28 AM
AVS Special Member
 
jdbimmer's Avatar
 
Join Date: Jan 2007
Posts: 1,003
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 14
Quote:
Originally Posted by anbjornk View Post

I believe the DTP94 is basically the same as the Display2... (Correct me if I'm wrong)

They are different. The DTP94 was a Monaco Systems developed product. My understanding is that Xrite/GMB bought Monaco, inherited and resold the DTP94 as "Xrite" and then discontinued it as it competed directly with their new D2 probes. It's a casualty of corporate mergers.
jdbimmer is offline  
post #22 of 315 Old 12-23-2007, 09:26 AM
Advanced Member
 
-Hitman-'s Avatar
 
Join Date: Nov 2004
Location: Manchester, UK
Posts: 626
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 20 Post(s)
Liked: 50
The DTP-94 also is better suited to Plasma display calibrations as the D2 picks up on the IR emissions, reporting an incorrect over-RED, Calman3 have added an error correction matrice to their software for the D2 to combat this error.
-Hitman- is online now  
post #23 of 315 Old 12-23-2007, 11:50 AM
AVS Special Member
 
derekjsmith's Avatar
 
Join Date: Oct 2003
Location: Mukilteo, WA
Posts: 1,890
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 98
Quote:
Originally Posted by jdbimmer View Post

They are different. The DTP94 was a Monaco Systems developed product. My understanding is that Xrite/GMB bought Monaco, inherited and resold the DTP94 as "Xrite" and then discontinued it as it competed directly with their new D2 probes. It's a casualty of corporate mergers.

Both the DTP94 and Display2 have a similar linage both were based on a Sequel Imaging reference design. The DTP94 was built by Moncao and the Display2 was by GMB. The hardware is very similar and the only real difference is the DTP94 has built in ambient temperature compensation to control temperature drift of the sensors. The main difference is the device driver and software interfaces are very different. When x-rite acquired GMB and Monaco they did not need two devices so close in design and capability so one of them got dropped the DTP94.

Derek

CTO / Founder - SpectraCal Inc.
derekjsmith is offline  
post #24 of 315 Old 12-23-2007, 11:56 AM
AVS Special Member
 
derekjsmith's Avatar
 
Join Date: Oct 2003
Location: Mukilteo, WA
Posts: 1,890
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 98
Quote:
Originally Posted by Neonmod View Post

The DTP-94 also is better suited to Plasma display calibrations as the D2 picks up on the IR emissions, reporting an incorrect over-RED, Calman3 have added an error correction matrice to their software for the D2 to combat this error.

Yes we do have a fix for the Display2 being used on Plasma but we are not going in to detail on how it was done, lets just say it was in cooperation with x-rite. We also have a fix for the Display2 when being used on a DLP with a fast strobe with it hanging, taking a long time to read or just timing out. The Display2 has some build in features that makes it harder to use on display types other than a direct view CRT or LCD which is what is was designed for.

Derek

CTO / Founder - SpectraCal Inc.
derekjsmith is offline  
post #25 of 315 Old 12-23-2007, 07:28 PM
Newbie
 
Ol' Knobby Joe's Avatar
 
Join Date: Nov 2007
Posts: 10
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
Can you comment on the accuracy of these meters when measuring an LCD screen with direct contact?

EDIT: Sorry, I forgot that you did post that information.
Ol' Knobby Joe is offline  
post #26 of 315 Old 12-26-2007, 01:57 PM
Member
 
Renovator's Avatar
 
Join Date: Oct 2007
Location: League City, Texas
Posts: 84
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 10
Quote:
Originally Posted by derekjsmith View Post

... The main difference is the device driver and software interfaces are very different. ...

My first rodeo.

I'll be learning on my plasma and smaller lcd monitors.

I've narrowed to an I-1 D2 or the DTP-94... but since I don't have a clue, I'm wondering about OS support.

Will the older DTP-94 driver or a required interface load under Vista ?

(I'm assuming CalMan does.)

Thanx
Renovator is offline  
post #27 of 315 Old 12-26-2007, 02:32 PM
AVS Special Member
 
derekjsmith's Avatar
 
Join Date: Oct 2003
Location: Mukilteo, WA
Posts: 1,890
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 98
Quote:
Originally Posted by Renovator View Post

My first rodeo.

I'll be learning on my plasma and smaller lcd monitors.

I've narrowed to an I-1 D2 or the DTP-94... but since I don't have a clue, I'm wondering about OS support.

Will the older DTP-94 driver or a required interface load under Vista ?

(I'm assuming CalMan does.)

Thanx

Yes CalMAN V3 supports Vista and XP. We also have a downloadable device driver pack for all the meters we support from the manufactures that also supports Vista and XP.

Derek

CTO / Founder - SpectraCal Inc.
derekjsmith is offline  
post #28 of 315 Old 12-26-2007, 06:30 PM
AVS Special Member
 
zoyd's Avatar
 
Join Date: Sep 2006
Location: Planet Dog
Posts: 4,427
Mentioned: 6 Post(s)
Tagged: 0 Thread(s)
Quoted: 75 Post(s)
Liked: 299
Quote:
Originally Posted by Neonmod View Post

The DTP-94 also is better suited to Plasma display calibrations as the D2 picks up on the IR emissions, reporting an incorrect over-RED, Calman3 have added an error correction matrice to their software for the D2 to combat this error.

This is not quite correct, the D2 blocks IR emission with an internal IR-cut filter just like other tristimulus probes. (This has been independently confirmed by both me and Bill from Calman using external filters, and by looking at the D2 patent). Also, as Tom's data points out, not all D2's have the excess red reading on plasma's compared to the i1pro. His old one does but his new one doesn't, I have one that does and the Calman team have at least one also. The problem is that the calibration matrix itself is not optimized for plasma. The CRT matrix may be slightly better than the LCD matrix but that has not been shown to be the case by anyone on this forum. The new HCFR version allows the probe to be run with the CRT matrix during plasma use if the user desires. I assume the new Calman allows for this also. Any other error correction matrix would not be advisable unless you are sure your probe has the same error, obviously Tom would not want to use such a matrix with his new D2.
zoyd is online now  
post #29 of 315 Old 12-26-2007, 07:59 PM
Advanced Member
 
ghibliss's Avatar
 
Join Date: Aug 1999
Location: New York, NY
Posts: 740
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 0 Post(s)
Liked: 11
Progressive Labs CA-6X software package has supported the D2 and i1Pro for some time now. We have since its introduction allowed the user the ability to select from the two calibration matrices stored in the eeprom (CRT/LCD) via our GUI main panel. We have never had the hanging issue which others have experienced as our code had the intelligence built in to deal with this issue from day one. We offer users of this instrument and more then eighteen others the opportunity to purchase our software at a reduced price.

For additional information contact us at our office anytime at 212-254-3541 or via email at progressivelabs@nyc.rr.com
ghibliss is offline  
post #30 of 315 Old 01-05-2008, 12:08 PM - Thread Starter
AVS Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 6,353
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 14 Post(s)
Liked: 204
Based on my tiny sample of 2 meters of each type, using plasma for the D2 and CRT for the Spyder2, it looks like the inter-meter variability is approximately:

D2: x 0.004, y0.008
Spyder2: x0.003, y0.027

As you can see, the y deviation is greater than x and truly profound on the Spyder2.

Quote:
Originally Posted by zoyd View Post

Tom, very nice set of measurements. Were the D2 and S2 probe variations (new vs. old) inside or outside of their factory uncertainties? (S2: +/- 0.002(LCD) +/- 0.0035(CRT) and D2:+/- 0.004 in x,y neglecting Y)


Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
Reply Display Calibration

User Tag List

Thread Tools
Show Printable Version Show Printable Version
Email this Page Email this Page


Forum Jump: 

Posting Rules  
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off