Yes, many thanks to Greg and Kevin for explaining these things to us!
I'm sorry that I haven't posted recently, as I've been plenty busy, but I have been reading and keeping up with the posts. I wonder if Jeff has been lurking as well and might consider adding CIE 1976 charts in a future release.
The next chart is from the same measurement session, and it will hopefully help to explain some of the readings in the CIE 1931 graph:
http://members.cox.net/snidely/postcaldata.jpg
Here we can clearly see that the readings I took below 30% stimulus are inaccurate. I believe this was due to the fact that my projector was mounted normally in its ceiling mount and the readings were taaken in emission mode off of the screen. AAs I can see it, there would have been two better ways to take these measurements that would have resulted in more accuracy in the low percentage stimuli:
1. Move the projector closer to the screen so that the projected image was MUCH smaller (like around 32"X18"), resulting in much higher ftL coming off of the screen and thus moving the measurements into an easier range for the EyeOne to measure.
2. Leave the projector where it was, use a tripod, put a diffuser on the EyeOne sensor (included in the Beamer package), and then take the measurements in the ambient mode, reading directly from the projector lens from a distance of about 2 feet. This would have a very similar effect of moving the measurements to a higher range like in solution #1, except that any color shift happening in the screen would not be considered. Since I am using a Firehawk, my belief is that this color shift would be neglible, but it is still less accurate (IMHO) than taking the measurements off of the screen.
Please also note that the inaccurate measurements in the low percent stimuli also throws any meaningful CR calculations right out the window, though it appears as if no harm was done to the gamma curve measurement, as no numbers were included in its calculations. I also assume that the color coordinates are being affected, as well as the obvious percentage stimulus errors. My intention is to remeasure in ambient mode from about 2 feet once my new 8.5' tripod arrives here (my current tripod is too short for this type of measuring).
Comments? Further clarifications of what we are seeing here?
I'm sorry that I haven't posted recently, as I've been plenty busy, but I have been reading and keeping up with the posts. I wonder if Jeff has been lurking as well and might consider adding CIE 1976 charts in a future release.
The next chart is from the same measurement session, and it will hopefully help to explain some of the readings in the CIE 1931 graph:
http://members.cox.net/snidely/postcaldata.jpg
Here we can clearly see that the readings I took below 30% stimulus are inaccurate. I believe this was due to the fact that my projector was mounted normally in its ceiling mount and the readings were taaken in emission mode off of the screen. AAs I can see it, there would have been two better ways to take these measurements that would have resulted in more accuracy in the low percentage stimuli:
1. Move the projector closer to the screen so that the projected image was MUCH smaller (like around 32"X18"), resulting in much higher ftL coming off of the screen and thus moving the measurements into an easier range for the EyeOne to measure.
2. Leave the projector where it was, use a tripod, put a diffuser on the EyeOne sensor (included in the Beamer package), and then take the measurements in the ambient mode, reading directly from the projector lens from a distance of about 2 feet. This would have a very similar effect of moving the measurements to a higher range like in solution #1, except that any color shift happening in the screen would not be considered. Since I am using a Firehawk, my belief is that this color shift would be neglible, but it is still less accurate (IMHO) than taking the measurements off of the screen.
Please also note that the inaccurate measurements in the low percent stimuli also throws any meaningful CR calculations right out the window, though it appears as if no harm was done to the gamma curve measurement, as no numbers were included in its calculations. I also assume that the color coordinates are being affected, as well as the obvious percentage stimulus errors. My intention is to remeasure in ambient mode from about 2 feet once my new 8.5' tripod arrives here (my current tripod is too short for this type of measuring).
Comments? Further clarifications of what we are seeing here?