Quote:
Originally Posted by giomania View Post
No go for me; the same results as before with both 12 and 21 points. Then I started examining the completed measurements, and I went through and corrected (re-measured) any color over dE 1, and got most of them under 1, with the exception of the first colors (red, green, blue).

So the only difference for the measurements that I was able to fix was that I manually re-touched (by pressing M) to get them under 1. These measurements finish the process slower (more accurately?) than Auto cal. Would inserting a wait time between measurements fix this?

The first colors (red, green, blue) are in the hundreds delta, and I cannot get them lower. This has to be the issue I am having; see below image.

No, those three colors don't matter. In fact you can skip them entirely. I am happy to continue troubleshooting this, but I must insist that you send ChromaPure support a calibration report and the log file from the auto-cal session. Otherwise, we could go back-and-forth forever on this. I must have the data that allows me to see exactly what's going on.



Inserting additional wait time couldn't hurt, but it will slow down the process.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
Sponsored Links
Advertisement
 
post #1712 of 1732 Old 07-14-2019, 03:54 AM - Thread Starter
AVS Forum Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 7,885
Mentioned: 21 Post(s)
Tagged: 0 Thread(s)
Quoted: 995 Post(s)
Liked: 837
Quote:
Originally Posted by CadH View Post
Any Chance for RPI support or definitely Not, Tom?
I would need the device and developer documentation describing how to build application support.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #1713 of 1732 Old 07-14-2019, 06:17 AM
QuadMersed in bass!
 
giomania's Avatar
 
Join Date: Nov 2001
Location: Chicago, Illinois
Posts: 3,842
Mentioned: 13 Post(s)
Tagged: 0 Thread(s)
Quoted: 621 Post(s)
Liked: 381
Quote:
Originally Posted by TomHuffman View Post
No, those three colors don't matter. In fact you can skip them entirely. I am happy to continue troubleshooting this, but I must insist that you send ChromaPure support a calibration report and the log file from the auto-cal session. Otherwise, we could go back-and-forth forever on this. I must have the data that allows me to see exactly what's going on.



Inserting additional wait time couldn't hurt, but it will slow down the process.
I agree we have done all that we can. I submitted the support request just now, and indicated that I have the files to submit. It did not allow for submission of files, so I presume that request is via email follow-up.

Mark
giomania is offline  
Sponsored Links
Advertisement
 
post #1714 of 1732 Old 07-14-2019, 10:22 AM - Thread Starter
AVS Forum Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 7,885
Mentioned: 21 Post(s)
Tagged: 0 Thread(s)
Quoted: 995 Post(s)
Liked: 837
Quote:
Originally Posted by giomania View Post
I agree we have done all that we can. I submitted the support request just now, and indicated that I have the files to submit. It did not allow for submission of files, so I presume that request is via email follow-up.

Mark
I don't know what "It did not allow for submission of files" means. Just send an email; to [email protected] and attach the requested files.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #1715 of 1732 Old 07-14-2019, 12:19 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
Tom,
Are the Display 3 Pros known to have bad 0% stimulus errors? Mine for example now has a Blue error of 249.5%, Red 30.1% & Green 105.7% with the sensor on a black surface like one does with the i1pro. I'm beginning to think that this is effecting my 5% stimulus when calibrating tv. It also seems to be getting worse over time. Is it fixable?
David

-Dave
DavidES is online now  
post #1716 of 1732 Old 07-14-2019, 12:32 PM
AVS Forum Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 4,075
Mentioned: 37 Post(s)
Tagged: 0 Thread(s)
Quoted: 1103 Post(s)
Liked: 1210
Quote:
Originally Posted by DavidES View Post
Tom,
Are the Display 3 Pros known to have bad 0% stimulus errors? Mine for example now has a Blue error of 249.5%, Red 30.1% & Green 105.7% with the sensor on a black surface like one does with the i1pro. I'm beginning to think that this is effecting my 5% stimulus when calibrating tv. It also seems to be getting worse over time. Is it fixable?
David
I'm not Tom, but if you mean the I1D3, it isn't meant to work that way. No black calibration is needed as it is for the I1 Pro. The I1 Pro's dark cal is done so the software can detect and adjust for noise in its readings over time. Different tech, different requirements. If your 5% is very dark, that can be a problem, as most low-cost tristim probes including the I1D3 struggle with color readings at lower light levels. Luminance can be read successfully lower than color can - down to .02 nits on the I1D3, IIRC.

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931
Rolls-Royce is online now  
post #1717 of 1732 Old 07-14-2019, 12:59 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
I've been trying to recalibrate my grayscale. My sensor is telling me that I have too much blue at 5% stimulus. I've run out of blue control trusting the sensor. I look at the screen there is no blue cast. Could the sensor circuitry be failing?

-Dave
DavidES is online now  
post #1718 of 1732 Old 07-14-2019, 01:19 PM
AVS Forum Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 4,075
Mentioned: 37 Post(s)
Tagged: 0 Thread(s)
Quoted: 1103 Post(s)
Liked: 1210
Quote:
Originally Posted by DavidES View Post
I've been trying to recalibrate my grayscale. My sensor is telling me that I have too much blue at 5% stimulus. I've run out of blue control trusting the sensor. I look at the screen there is no blue cast. Could the sensor circuitry be failing?
What is your luminance reading at 5%? Again, the meter will have problems accurately reporting color as luminance gets close to its lower sensitivity limit. It's a known issue with lower-cost colorimeters.

What display do you have?

How old is your meter?

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931
Rolls-Royce is online now  
post #1719 of 1732 Old 07-14-2019, 01:36 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
Quote:
Originally Posted by Rolls-Royce View Post
What is your luminance reading at 5%? Again, the meter will have problems accurately reporting color as luminance gets close to its lower sensitivity limit. It's a known issue with lower-cost colorimeters.

What display do you have?

How old is your meter?

I don't remember. I do remember around 99.3% of 2.3 gamma if that helps. I didn't save the session.


TCL 49S405


October 2018

-Dave
DavidES is online now  
post #1720 of 1732 Old 07-14-2019, 01:40 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
I also remember it saying 124% Blue at the 5% stimulus.

-Dave
DavidES is online now  
post #1721 of 1732 Old 07-14-2019, 02:04 PM
AVS Forum Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 4,075
Mentioned: 37 Post(s)
Tagged: 0 Thread(s)
Quoted: 1103 Post(s)
Liked: 1210
Quote:
Originally Posted by DavidES View Post
I don't remember. I do remember around 99.3% of 2.3 gamma if that helps. I didn't save the session.


TCL 49S405


October 2018
The meter is less than a year old. It's possible but not likely that it has drifted.

What I'm looking for is the actual nits or foot-lambert reading you got at that 5% stimulus. A percentage of gamma tells me nothing, since the calculated gamma target is based on measurement of full white and full black.

I do suspect that your luminance at 5% stimulus may be too low for the meter to accurately report colors. Next time, try a continuous read at that 5% mark. If red, green, or blue are fluctuating a lot there, it's too little light for accuracy. Keep in mind that in Rec 709 (SDR HD), Blue makes up only 7.22% of the luminance of D65. It's the dimmest color, so is toughest for the meter to read, especially at low stimulus levels.

See post number 15 here:https://www.avsforum.com/forum/139-d...e-balance.html

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931

Last edited by Rolls-Royce; 07-14-2019 at 02:09 PM.
Rolls-Royce is online now  
post #1722 of 1732 Old 07-14-2019, 05:15 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
Quote:
If red, green, or blue are fluctuating a lot there, it's too little light for accuracy.

What amount of fluctuation would be considered a lot? I do get some; between 2-4%.



I remeasured in the White Balance and got at 5% Y 0.106 - 0.108. Red and Green are between 97 -99%. Blue at around 127%.

-Dave
DavidES is online now  
post #1723 of 1732 Old 07-14-2019, 06:21 PM
AVS Forum Special Member
 
Rolls-Royce's Avatar
 
Join Date: Apr 2003
Location: Victorville, CA
Posts: 4,075
Mentioned: 37 Post(s)
Tagged: 0 Thread(s)
Quoted: 1103 Post(s)
Liked: 1210
Quote:
Originally Posted by DavidES View Post
What amount of fluctuation would be considered a lot? I do get some; between 2-4%.



I remeasured in the White Balance and got at 5% Y 0.106 - 0.108. Red and Green are between 97 -99%. Blue at around 127%.
Based on those numbers, Blue is at ~ 0.009719564‬ to 0.009902952‬ nits. Probably way too low to be read accurately: it's less than half of the I1D3 minimum luminance.

...Royce...

"I never drink...wine."
Bela Lugosi, DRACULA, 1931
Rolls-Royce is online now  
post #1724 of 1732 Old 07-14-2019, 06:24 PM
QuadMersed in bass!
 
giomania's Avatar
 
Join Date: Nov 2001
Location: Chicago, Illinois
Posts: 3,842
Mentioned: 13 Post(s)
Tagged: 0 Thread(s)
Quoted: 621 Post(s)
Liked: 381
Quote:
Originally Posted by Rolls-Royce View Post
The meter is less than a year old. It's possible but not likely that it has drifted.



What I'm looking for is the actual nits or foot-lambert reading you got at that 5% stimulus. A percentage of gamma tells me nothing, since the calculated gamma target is based on measurement of full white and full black.



I do suspect that your luminance at 5% stimulus may be too low for the meter to accurately report colors. Next time, try a continuous read at that 5% mark. If red, green, or blue are fluctuating a lot there, it's too little light for accuracy. Keep in mind that in Rec 709 (SDR HD), Blue makes up only 7.22% of the luminance of D65. It's the dimmest color, so is toughest for the meter to read, especially at low stimulus levels.



See post number 15 here:https://www.avsforum.com/forum/139-d...e-balance.html


In the post I made above, you can see that at 5%, the luminance was 4.02 with a delta of 18.62. I had to try manual measurements a few times to get it that low using an i1 Display Pro 3.

Mark


Sent from my iPhone using Tapatalk Pro
giomania is offline  
post #1725 of 1732 Old 07-14-2019, 06:37 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
Quote:
In the post I made above, you can see that at 5%, the luminance was 4.02 with a delta of 18.62. I had to try manual measurements a few times to get it that low using an i1 Display Pro 3.
4.02 is your gamma / EOTF, the r symbol. Your luminance Y is in the top line, third number 0.0595 nits. The top line is xyY readings.
giomania likes this.

-Dave

Last edited by DavidES; 07-14-2019 at 06:39 PM. Reason: clarification
DavidES is online now  
post #1726 of 1732 Old 07-14-2019, 06:41 PM
Member
 
Join Date: Nov 2018
Location: Chattanooga, TN
Posts: 15
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 8 Post(s)
Liked: 2
Quote:
Originally Posted by Rolls-Royce View Post
Based on those numbers, Blue is at ~ 0.009719564‬ to 0.009902952‬ nits. Probably way too low to be read accurately: it's less than half of the I1D3 minimum luminance.


Thanks, Rolls-Royce.
Rolls-Royce likes this.

-Dave
DavidES is online now  
post #1727 of 1732 Old 07-18-2019, 08:00 PM - Thread Starter
AVS Forum Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 7,885
Mentioned: 21 Post(s)
Tagged: 0 Thread(s)
Quoted: 995 Post(s)
Liked: 837
Quote:
Originally Posted by CadH View Post
Clearing my recent browser data got rid of the problem for me with Opera, maybe Safari is the same.

Greetings,
Cornelius
FYI. On June 28 our web site was hacked. I didn't notice it because when you enter the URL directly it works fine. The problem only appears when you did a web search on ChromaPure. The search would appear to yield the correct site, but when you clicked on the link it would take you to an online pharmacy.


I have now fixed the problem. However, you may have to clear your browser's cache.
namirkhan likes this.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #1728 of 1732 Old 07-26-2019, 02:56 PM - Thread Starter
AVS Forum Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 7,885
Mentioned: 21 Post(s)
Tagged: 0 Thread(s)
Quoted: 995 Post(s)
Liked: 837
Lumagen Auto-cal fix

I have now uploaded a version of the 3.1.19 Preview Release that includes two changes:
1. Adds support for auto-cal along with the SpyderX.
2. Adds a fix for the auto-cal module that resolves the problem that appears when selecting different grayscale modes. I simply removed the mode selector from the auto-cal module. Now, auto-cal uses whichever grayscale mode you have selected in the main Options module.


I tested this fix for both 12 point and 21 point modes and saw no problems. Here's the result for a 21 point HDR/2020 calibration.




giomania likes this.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #1729 of 1732 Old 08-11-2019, 10:46 PM
Member
 
rachmat-d's Avatar
 
Join Date: Jul 2006
Posts: 164
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 56 Post(s)
Liked: 14
Tom,

For HDR calibration, if i using auto cal why the grayscale is not good ?
Attached Thumbnails
Click image for larger version

Name:	BEFORE.jpg
Views:	25
Size:	69.3 KB
ID:	2602028   Click image for larger version

Name:	AFTER.jpg
Views:	25
Size:	63.5 KB
ID:	2602030  
rachmat-d is offline  
post #1730 of 1732 Old 08-11-2019, 10:57 PM - Thread Starter
AVS Forum Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 7,885
Mentioned: 21 Post(s)
Tagged: 0 Thread(s)
Quoted: 995 Post(s)
Liked: 837
Lumagen has told me that the color matrix calibration can throw the grayscale off some, so it is a good idea to do a grayscale touch-up after color calibration has finished.


In your case it looks like to need to edit three or four points.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
post #1731 of 1732 Old 08-12-2019, 12:15 AM
Member
 
rachmat-d's Avatar
 
Join Date: Jul 2006
Posts: 164
Mentioned: 0 Post(s)
Tagged: 0 Thread(s)
Quoted: 56 Post(s)
Liked: 14
i am doing autocal without grayscale and then after that i do manual grayscale , it is ok ?
How do i verify the color , as the number of color show around 200-500 something ?
rachmat-d is offline  
post #1732 of 1732 Old 08-12-2019, 02:44 AM - Thread Starter
AVS Forum Special Member
 
TomHuffman's Avatar
 
Join Date: Jan 2003
Location: Springfield, MO
Posts: 7,885
Mentioned: 21 Post(s)
Tagged: 0 Thread(s)
Quoted: 995 Post(s)
Liked: 837
Quote:
Originally Posted by rachmat-d View Post
i am doing autocal without grayscale and then after that i do manual grayscale , it is ok ?
How do i verify the color , as the number of color show around 200-500 something ?
No, do grayscale first and then after color, spot correct grayscale as needed.

You verify by the report that you are prompted to run at the end.

Tom Huffman
ChromaPure Software/AccuPel Video Signal Generators
ISF/THX Calibrations
Springfield, MO

TomHuffman is offline  
Sponsored Links
Advertisement
 
Reply Display Calibration

Thread Tools
Show Printable Version Show Printable Version
Email this Page Email this Page


Forum Jump: 

Posting Rules  
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are Off
Pingbacks are Off
Refbacks are Off