i1Pro low "Y" reading values fatal?

393 Views 3 Replies 3 Participants Last post by  nw06
Apologies up front, as I do not know the technical details behind calibration, so some of my following terminology may be incorrect.

I have i1Pro RevD + i1DisplayPro meters that have been in storage for about 5 years, since I last calibrated my plasma tv. I recently got them back out to attempt to redo some of my calibrations, however, I have noticed that the i1Pro gives significantly lower "Y" reading values than the i1DisplayPro. For example, I know my VT60 is hitting about 115nits white peak, which the I1DisplayPro agrees with, but the i1Pro says 85nits...
I did read somewhere that i1Pro's do lose that light sensitivity as they age (mine was second hand when I bought it in the first place).

My question, is the i1Pro effectively useless for calibration/correction of i1Display now?
The RGB white balance graphs are fairly similar for both meters, but obviously gamma looks different between them.
Is the Y reading independent of the RGB readings? i.e. make a correction matrix to i1Display to use i1Pro RGB readings, but keep it's native Y reading??
Sorry if this makes no sense.
1 - 4 of 4 Posts
Apologies up front, as I do not know the technical details behind calibration, so some of my following terminology may be incorrect.

I have i1Pro RevD + i1DisplayPro meters that have been in storage for about 5 years, since I last calibrated my plasma tv. I recently got them back out to attempt to redo some of my calibrations, however, I have noticed that the i1Pro gives significantly lower "Y" reading values than the i1DisplayPro. For example, I know my VT60 is hitting about 115nits white peak, which the I1DisplayPro agrees with, but the i1Pro says 85nits...
I did read somewhere that i1Pro's do lose that light sensitivity as they age (mine was second hand when I bought it in the first place).

My question, is the i1Pro effectively useless for calibration/correction of i1Display now?
The RGB white balance graphs are fairly similar for both meters, but obviously gamma looks different between them.
Is the Y reading independent of the RGB readings? i.e. make a correction matrix to i1Display to use i1Pro RGB readings, but keep it's native Y reading??
Sorry if this makes no sense.
Hi, there capability to send your i1PRO to X-Rite for re-calibration, but it's expensive service.

When you create a 4 color matrix meter correction table, if you create the 3x3 matrix manually and enter '1' numbers to the 8x Y numbers (4x Y of WRGB of i1Display PRO and 4x Y of WRGB of i1PRO), this will not correct the Y difference your meters have.
I have i1Pro RevD + i1DisplayPro meters that have been in storage for about 5 years, since I last calibrated my plasma tv. I recently got them back out to attempt to redo some of my calibrations, however, I have noticed that the i1Pro gives significantly lower "Y" reading values than the i1DisplayPro. For example, I know my VT60 is hitting about 115nits white peak, which the I1DisplayPro agrees with, but the i1Pro says 85nits...
In my experience it's not unusual for different instruments to have noticeably different absolute Y calibration differences, even when brand new from the factory. (I think X-Rite may have improved the correspondence between the i1D3 and i1Pro2 with recent products though.). This discrepancy isn't really that significant in the scheme of things, since it's all about the color and luminance gradation. An ArgyllCMS/DisplayCAL correction matrix will make the two instruments have corresponding Y values anyway.

You might want to clean the filter on the i1Pro2 though, just to make sure it doesn't have dust on it.
In my experience it's not unusual for different instruments to have noticeably different absolute Y calibration differences, even when brand new from the factory. (I think X-Rite may have improved the correspondence between the i1D3 and i1Pro2 with recent products though.). This discrepancy isn't really that significant in the scheme of things, since it's all about the color and luminance gradation. An ArgyllCMS/DisplayCAL correction matrix will make the two instruments have corresponding Y values anyway.

You might want to clean the filter on the i1Pro2 though, just to make sure it doesn't have dust on it.
Thanks Ted & gwgill

So, lower Y values than actual does not impact the colour accuracy of the i1Pro?
I just have to remember that the actual Y reading is likely about 25-30% higher than i1Pro is saying, and set peak brightness accordingly?

I have been using HCFR/DisplayCal, and the matrix correction does indeed make the i1D3 read exactly the same values as i1Pro...I was just not sure if that was ok given I knew the Y values were low (I of course have no idea if the xy values are more accurate then the i1D3 either of course, but assuming so).

Thanks for the help!